Posted

Here is this Stowa balance assembly. The outer coul of the spring is resting on the internal regulator pin if i set the regulator to good rate in DU and DD positions. It has ~30-40 sec/day gain in PU and the same loss in PD. If i set it to lower rate then the outer coil barely leaving the regulator pin. It should be between them at rest. 

If i remove the spring from the balance and mount it on the cock then the collet goes to the point marked with the red circle.

Where and how much to manipulate the spring? I was watching Marks videos and reading the related part in De Carle watch adjusting book but still not get it. I dont want to ruin the watch after working so much with it. Please help!

 

Snapshot_20160928_1_2.jpg

Posted

Can you re-pin the hs on the end stud to be an hair longer? That could avoid manipulation of the dog leg which is always a risky proposition.


  • Like 1
Posted
18 minutes ago, clockboy said:

Thank You! Investigating the description. 

12 minutes ago, jdm said:

Can you re-pin the hs on the end stud to be an hair longer? That could avoid manipulation of the dog leg which is always a risky proposition.

Yes, i was thinking about that and excluded as a solution. If i make the HS longer then i need to move the regulator further to keep time and will end up in the smae situation. 

Posted

The point of having a correct overall hs length, as well the smaller beat error possible is to keep the regulator in the middle of its range. That and a good hs shape, should make so all is good. How is the timegrapher picture?

Posted
3 minutes ago, jdm said:

 How is the timegrapher picture?

This was at the beginning:

Working.jpg

Then i noticed, that the stud is mounted to low and the HS was not flat. Moreover, the regulator pin was bent. (I think it has to much distance between the pins after reading the description linked by Clockboy. I will reshape that to get hem closer.). Corrected the flatness and straightened the pin and it got better amplitude. Both pictures are dial down, the same is with dial up. But terrible performance if the movement is sideways.

Working after hairspring tinkering.jpg

Posted

The regulator pin distance is relative to how much the hs is "breathing", again this depends on its shape and amplitude. Personally I woud try a small manipulation of the hs to contrast the sagging when vertical, rather than messing with the regulator pins. But in this case you are doing positional adjustment and that is more an art than science.


  • Like 1
Posted

No matter where the regulator is the hair spring should be able to bounce between the regulator pins and no two coils of hair spring should be touching. 

  • Like 1
Posted

Ok, the description above and oldhippy saying that the most important is that the HS ens curve stays perfectly centered between the pins no matter the regulator position. Then i will need to take off the HS and first make sure of this. And then? The collet position can be set by the dog leg 2x 45 degrees bending? And how can i determine the distance between the two bendings? Or just make the bendings and change the degree of them so that the collet get over the jewel?

Posted

I wouldn't alter the dog legs. All you have to do is to correct slightly the final curve. If you look on youtube there is a video by perplxr doing that on a running movement.


  • Like 1
Posted

Thank You! I am checking the video. But i am sure that the collet will move even closer to the pins if i bend a bit on the end curve. I will check that at home. 

Posted

To address the collet being off center, you can try the followoing

a). Manipulate the spring at point A to move the collet in an arc so that it gets centered on the cock.

b). Manipulate point B so that the whole hairspring shifts in the direction of the arrow.

Also check that the spring run from the stud to regulator is consistent, ie if it is in the centre when the regulator is centred, it should be centred when regulator is set to fast or slow.

Enjoy yourself.

Anil

 

  • Like 1
Posted

Mark has some good videos about hairsprings. You could type into youtube search watch hairsprings and Marks videos will come up including many others. It is best to watch the videos then for me to explain what you should do.

  • Like 1
Posted
7 minutes ago, anilv said:

To address the collet being off center, you can try the followoing

a). Manipulate the spring at point A to move the collet in an arc so that it gets centered on the cock.

b). Manipulate point B so that the whole hairspring shifts in the direction of the arrow.

Also check that the spring run from the stud to regulator is consistent, ie if it is in the centre when the regulator is centred, it should be centred when regulator is set to fast or slow.

Enjoy yourself.

Anil

 

Thank You! Please show me the references for points A and B :)

That perplxr guy omit showing the most important manipulation. At 10:25 one can clearly see that the end curve is following the pins. Especially at the stud. And at 11:00 he is showing the already repaired spring. He must have done soemthing at the stud mounting point. 

Posted
1 hour ago, szbalogh said:

Thank You! Please show me the references for points A and B :)

That perplxr guy omit showing the most important manipulation. At 10:25 one can clearly see that the end curve is following the pins. Especially at the stud. And at 11:00 he is showing the already repaired spring. He must have done soemthing at the stud mounting point. 

I don't know. There is always a bit of assumption and uncovered areas in their videos. Maybe they keep the complete information for the actual students only :) 

Posted
2 minutes ago, jdm said:

I don't know. There is always a bit of assumption and uncovered areas in their videos. Maybe they keep the complete information for the actual students only :) 

"At 10:25 one can clearly see that the end curve is NOT following the pins. " I also omitted the most important word :)

Watching Marks video again and again and just realized that his patient has a bending at the stud. It seems that i need to move the regulator next to the stud, make the start of the curve between the pins and then form the proper arc from there.

Shortly, i need to start shaping from the stud. 

Now all is clear, but to much coffe for today... lets hope i wont make it worse :)

  • Like 1
Posted

Here we go. Internal regulator pin pushing hard on the HS.

Snapshot117.jpg

End curve shaped. At slowest position:

Snapshot121.jpg

And in fastest position:

Snapshot122.jpg

The collet was of course way to far from center. The dog leg was made.

Snapshot123.jpg

Tried a bigger bend... hell nooo

Snapshot124.jpg

With sharper turns:

Snapshot125.jpg

Then i noticed that small bending right after the dog leg in the outer coil. Corrected and looking almost good:

Snapshot127.jpgSnapshot128.jpg

Currently it is looking like this: 

Snapshot129.jpg

What do you think? Where should i do more corrections?

Snapshot126.jpg

Snapshot120.jpg

Posted (edited)

Just a thought but I always get suspicious about the point the HS exits the stud. I have sorted out more out of flat and off centre hair springs by tweaking at the stud than anywhere else and I reckon it's because it's the most vulnerable point if the balance cock is handled carelessly or if there is a struggle with the tweezers trying to remove the stud from the mounting.

Looking at your pics it looks to me like the terminal curve spacing (to the next coil) is less at the stud than at the dog leg.Image4.jpg

The difference between the red and green circles.

If you slacken the bends at the dog leg it will reduce the spacing at the dog leg, and if you slacken the bends at and just after the stud it will increase the spacing at the stud; in either case you will need to gently reprofile the arc between the two points. Also, it looks as though both approaches should be possible without adversely displacing the collet from the balance pivot, and could even be used to improve this alignment.

What we can't see though is the where the central point between the inner curb pin and the inside of the boot is. As this is the point you need the hair spring to be at, without knowing where that point is it is impossible to say which of the two adjustments is correct, or if you need to do a bit of both, which I suspect is the correct answer.

If you measure the distance between the centre of the balance jewel and the inside of the inner curb pin and carefully draw a circle on a piece of paper with that diameter using a pair of compasses. Place this template onto a piece of cork board and pin the hair spring to it so that the pin goes into the centre of the circle you will be able to use the circle as a template for the terminal curve. You do need to ensure that the collet is held firmly on the centre of the circle.

Edited by Marc
  • Like 3
Posted
8 hours ago, Marc said:

Just a thought ...

Thank You for the time and help!

Yes, that end curve has a strange arc but surprisingly following the pin. 

Reassembled this morning and set the regulator to ok time in DD position. It is almost out of regulating space. Should i shorten the HS at the stud? That would ruin all of my work :( Or remove weight from the balance? The staff was repivoted both end so it is possible that weighing more now. 

Seems to me that the regulating pins are to far from each other. Or is it ok? As You suggested, first we should come to an agreement with this to make final shaping. 

It has nice amplitude now in DD even with much bigger beat error (placed back the stud where it was befor tinkering and have not corrected on the beat yet). But total scrap results in PU, PD and DU. Definitively dirt, since it shows irregular pattern. I will check this after proper cleaning (I had mounted and dismounted the separate parts many times).

Capture_20160930.wmv

DD after HS adjustment.jpgPD after HS adjustment.jpg

Posted

Ok, now i let the watch run for almost two days and here we are:

DD

DD after HS adjustment.jpg

DU

DU after HS adjustment.jpg

The repivoted pinions were polished to perfect length, no endshake. This is what i expected :)

But here comes the complication:

PU

PU after HS adjustment.jpg

PD

PD after HS adjustment.jpg

And PL

PL after HS adjustment.jpg

What are the experts saying to this? :)

  • Recently Browsing

    • No registered users viewing this page.
  • Topics

  • Posts

    • If you see at least one thread protruding, or even a half, then you may have a chance using this "tool", but don't squeeze hard. Lot's of patience, and if you notice any signs of loctite or other adhesives, then apply several drops of acetone on top of broken stem. Use heat (hairdrier) also. You should be able gradually turn it counterclockwise, just like i did on this broken Timex stem, which was glued with something like nail hardener. Then you can use an extender to save your broken stem. https://m.media-amazon.com/images/I/513DVvkfW5L.__AC_SX300_SY300_QL70_ML2_.jpg   You can also try cutting a small grove / slot on top of the broken stem to use 0.7mm flat screwdriver.   Depending on the Crown material and Stem material, you can try dissolving the stem. 
    • How long had it been without any wind before you first put it on the timegrapher?  Perhaps there was some debris / old lubricant that loosened up while you were waiting to do the 24 hour reading? I recently serviced a Seiko 6319 that immediately after service was absolutely horrible on the timegrapher - 24 hours later (without me doing anything) it was rock solid!  I was a little surprised by this (as I have seen improvements in amplitude after letting it run in but never a watch going from useless to great!). I was suspicious that some dirt must still be in there so I took it apart and cleaned and lubcricated it again and had exactly the same experience - snow storm that after some hours turned into a nice straigh line with good amplitude 🤷‍♂️
    • Yes, I know it was able to run before just enough that everything supposed to be affixed with shellac could fall out. I won't get back to work on this watch for several days at least, but one thing you mentioned about the banking pin adjustments being intended for fork horn clearance got me wondering.... The last adjustment I made to the entry stone yesterday did not solve the problem when the balance is installed. The escape teeth can't quite clear the stone when the balance swings. But I thought it was going to work because manually snapping the fork back and forth did allow the escape wheel to advance almost all the time. I wonder if there's another issue with the impulse jewel, that I have to shellac again anyway. I am going to measure the fork slot and the jewel because I am curious if it is the right size or possibly narrower than it should be. My thinking is if the horn clearance is too great maybe it is not getting enough impulse to clear the escape tooth. But pushing the fork with a tool can move it far enough to clear the escape wheel tooth.  I'll see later in the week. I didn't expect a big challenge from this watch because it ticked for 5 minutes before stopping without shellac. As always I was way wrong.
    • I had one success (and one failure) dissolving a broken stem out of a crown with Alum. The successes was a stainless steel crown and a milder steel stem. A Tag Heuer 2000 Automatic - here with the solution at around 60ºC I could see small bubbles coming up from the stump of the stem and it took maybe 6 hours to disolve I tried to repeat this with another watch and it did nothing.....  This was A Bulova Ambassador with what again looked like a stainless steel crown.  
    • That good to know in one way.
×
×