Hello and good day everyone.  I'm still getting my feet wet with watch repair, and would appreciate your advice regarding some timegrapher readings of a service I completed yesterday.

I bought an older Jerome Piquot watch off of eBay.  I know nothing about the brand, I just thought the watch looked unique and thought it'd be fun to rebuild.  Considering I paid under $25 for it, I figured if I botched it completely it wouldn't turn out to be an expensive way to learn a lesson in watch repair.   It has an AS 1686 movement.  The service went really well yesterday and it has been running in for around 9 hours at this point.  I'm getting a beat error of 0.8ms.  This AS 1686 does not have an adjustable beat error stud, I know that I would have to remove the balance wheel and adjust it manually if I wanted to.  I'd like to avoid that if possible and just wanted to call upon the experience of here to see if 0.8ms is something they'd try to improve upon on this watch.

Here's dial up running at +5 to +6 s/d, and has a very strong amplitude considering I didn't replace the mainspring, but I was meticulous with the cleaning and did polish the pivots with my recently acquired Jacot tool and remove a bit of extra side shake in the mainspring barrel.

image.png.946f07e2cb0640ef94963305909b0a7c.png

 

Here's dial down - running at -2 to -3 s/d with a slight amplitude drop from the previous position.

image.png.35a2e4754a1419c3c38af1ec734276a3.png

 

I feel pretty good about the service, and was encouraged to see single digits after a short period of regulation.  Would you bother with the 0.8ms beat error in this 60 year old movement?  Also, does the amplitude drop of 21 degrees between dial up and dial down seem within reason to you?  There is a bit more end shake in the 4th wheel than I'd ultimately like, but it is not so much that it would cause any part of the wheels to not be in contact with one another.  It just meets the escape wheel not in the center of it's pinion, but it is still in full contact throughout all end play testing between both the wheels.  As of right now though, I do not have a jeweling/Seitz tool needed to adjust for that.  I think it is as good as I can get it with my tools & skill level.  I'd appreciate any feedback you may have.  Thanks.

 

  • Like 1
Link to comment
Share on other sites

Companies I've done work for on both new builds and vintage repair stipulated a beat error under 1ms on watches with fixed stud holders (in 6 positions, haha). I recall in some Vibrograf literature, written before the advent of digital timing machines, they said anything under 2ms will have essentially zero effect on the running. So you look fine there!

  • Like 2
Link to comment
Share on other sites

1 minute ago, nickelsilver said:

Companies I've done work for on both new builds and vintage repair stipulated a beat error under 1ms on watches with fixed stud holders (in 6 positions, haha). I recall in some Vibrograf literature, written before the advent of digital timing machines, they said anything under 2ms will have essentially zero effect on the running. So you look fine there!

Good information to know.  Thank you!

Link to comment
Share on other sites

I would say that .8ms, your worrying about nothing. 

On the other hand, 315 degrees, especially with a used mainspring,  sounds a bit high.  Are you sure that you have the correct lift angle entered?

I do not think that it would have any effect on time keeping but it would be nice to know if the information it was providing was accurate.

Good luck. 

Shane 

Edited by Shane
Link to comment
Share on other sites

3 hours ago, thor447 said:

Also, does the amplitude drop of 21 degrees between dial up and dial down seem within reason to you? 

I suppose, your pivot polishing caused that difference. If a pivot end of the balance staff is falsely too high domed, this will happen in the corresponding flat position. How is your amplitude in vertical positions?

3 hours ago, thor447 said:

and just wanted to call upon the experience of here to see if 0.8ms is something they'd try to improve upon on this watch.

If you want to know, if xx ms is ok or not, that number does not yet tell too much (0.8 however will be ok in any case).
Essential is, how many degrees out of the middle is the balance in rest? That is the reason of the beat error.
The very same error of 5 degrees will show:
1,3 ms @ 28800, 300° Ampl. (B.E.= 5°)
2,1 ms @ 18000, 300° Ampl. (B.E.= 5°)
3,6 ms @ 18000, 180° Ampl. (B.E.= 5°)

That is why I regard the beat error in degrees only, which my PCTM shows. 2 degrees and lower will be near to perfect. Your 0.8 ms corresponds to 2.0 degrees!

Of course you cannot expect such sophisticated measurements from Weishis and Witschis 😉

Frank

  • Like 2
Link to comment
Share on other sites

5 hours ago, nickelsilver said:

Companies I've done work for on both new builds and vintage repair stipulated a beat error under 1ms on watches with fixed stud holders (in 6 positions, haha). I recall in some Vibrograf literature, written before the advent of digital timing machines, they said anything under 2ms will have essentially zero effect on the running. So you look fine there!

For watches with fixed stud holders, 2ms is my "reference point" as to whether to adjust or not. 

I think people get too fixated with beat error.

BTW the correct lift angle for the AS 1686 is 52° - which is the default on that timegrapher.

Link to comment
Share on other sites

Rolex is another interesting company for timing specifications. All sorts of timing specifications fully wound up half wound even have paperwork you're supposed to fill out but when it comes to the beat error that's pretty simple. A maximum of 0.8 ms test in any one of the five positions they normally test.

 

 

Link to comment
Share on other sites

11 hours ago, nickelsilver said:
5 hours ago, mikepilk said:

they said anything under 2ms will have essentially zero effect on the running. So you look fine there!

 

I must be being very hard on myself in this case,anything over 0.3 and I'm starting to feel agitation .

8 hours ago, praezis said:

I suppose, your pivot polishing caused that difference. If a pivot end of the balance staff is falsely too high domed, this will happen in the corresponding flat position. How is your amplitude in vertical positions?

If you want to know, if xx ms is ok or not, that number does not yet tell too much (0.8 however will be ok in any case).
Essential is, how many degrees out of the middle is the balance in rest? That is the reason of the beat error.
The very same error of 5 degrees will show:
1,3 ms @ 28800, 300° Ampl. (B.E.= 5°)
2,1 ms @ 18000, 300° Ampl. (B.E.= 5°)
3,6 ms @ 18000, 180° Ampl. (B.E.= 5°)

That is why I regard the beat error in degrees only, which my PCTM shows. 2 degrees and lower will be near to perfect. Your 0.8 ms corresponds to 2.0 degrees!

Of course you cannot expect such sophisticated measurements from Weishis and Witschis 😉

Frank

I'm  thinking power reserve would be a little less with a higher beat error than acceptable. The balance wheel would want to come to rest at a sooner time period ?

49 minutes ago, JohnR725 said:

Rolex is another interesting company for timing specifications. All sorts of timing specifications fully wound up half wound even have paperwork you're supposed to fill out but when it comes to the beat error that's pretty simple. A maximum of 0.8 ms test in any one of the five positions they normally test.

 

 

Five positions ?  Do they leave out DD ?

Link to comment
Share on other sites

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.
Note: Your post will require moderator approval before it will be visible.

Guest
Reply to this topic...

×   Pasted as rich text.   Restore formatting

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.



  • Recently Browsing

    • No registered users viewing this page.
  • Topics

  • Posts

    • I used to just remove the stem pop the movement in its case then put the stem back in check everything then back on the timer. 
    • That's the right technique Mal 👍 If you are ok with a loupe ( which I'm not ) then pick up the dial in your left hand and make a tripod holder out of 3 fingers and then exactly how Mal and me described with your right hand, press and flick out the stem at the same time. You need three fingers to do this, i use my ring ring finger , sounds like mal uses his little finger, just whatever you feel comfortable with. The key point and dangerous side to this is the tweezers or driver slipping off the release screw. Stablise your left arm by resting your elbow on your bench, so that you can hold the movement horizontally flat, use a x5 loupe to view  and good light so you can see well and have a good tight fitting screwdriver to push the release down. Or as suggested a pusher mounted solid upside-down somewhere then all you need to do is push your movement up to it. I'll rig something up in a bit to show you what i mean. Something like this, but tbh its better to just learn to do with your fingers.
    • Thanks Michael, shes a real soldier, it would take a tank to knock her off her feet.  Shes surprised even me and i knew she was a tough cookie.  Good for you Michael, you keep at it. I've been reading your posts re. your watchroom they're inspiring and it doesn't matter how long it takes you to get there its about the journey. Rome wasn't built in a day, our watchcaves are our solitudes, our own little empires where we can shut ourselves away from the crap that goes on around us in the world. I look forward to seeing you progress, I have my own ideas how i like to do things. But in answer to your questions on that thread, just let your workspace evolve around you and how you work. You wont figure it out straight away but you will know when to change it and when to leave it alone. But most of all just enjoy it.
    • That can be tricky. Sometimes I hold the movement from the edges in left hand and with right hand I push the release with tweezers and simultaneously pull the stem out with right hand pinky or something. 
    • I see what you mean. I must admit I’m not entirely comfortable with that but I can’t see a better method than what you suggest. I’d like to see something like this mirror with a short spike: https://www.cousinsuk.com/product/mirror-for-observing-movements Place it on, push the spike on the post with one hand and release the stem with the other hand. Minor problem, it doesn’t exist.
×
×