Smiths Everest Cal 01 04 timegrapher - why such different horizontal and vertical rates?


Recommended Posts

This has got me puzzled. This Smiths cal 01 04 is running strongly, but the vertical rates are all slow, and all over the place.

So I thought I'd do some dynamic balancing. The pic is from my balancing spreadsheet.
You can see (fully wound) on the left good amplitude in all position, and DU +5 and DD +9s OK. But all vertical rates slow.

With 57s between Crown up (-32s) and crown down (-89s), and slow in vertical positions, I was expecting a heavy spot on the balance.
But with 160° amplitude to determine the heavy spot (see pic on right), all the numbers are similar, suggesting no heavy spot 🥴

I demagnetised. If coils were sticking when vertical, you would expect a higher rate.
The hairspring looks to be 'breathing' correctly.

Usually when vertical rates are a little slower, I suspect too much gap at the regulator pins. But this is a lot, and the pins look OK to me.

What do you think?

 

image.png.1e89211ef546cb1335ec012ba95b28d4.png

 

 

 

 

Link to comment
Share on other sites

Two things, perhaps one more important then the other.

As far as I know, dynamic poising should only be done if a watch runs good. A quote from the "How to" series Dynamic poising;

"If a watch runs well—it is clean, free of flaws, and running with good amplitude—it can usually be adjusted to be more accurate across different positions".

https://adjustingvintagewatches.com/category/dynamic-poising/

If in vertical positions a watch runs only between 140-180, that's IMHO too low for dynamic poising. Also, at those lower (140 degrees) amplitudes, one has to question the reliability of the timegrapher readings (At least with my Chinese Weishi 1000).

The other, but a very minor point, and that doesn't explain the low vertical position amplitudes, is the hairspring breathing in the regulator pin/boot. It seems that the hairspring barely loosens itself from the regulator pin. It doesn't seem touching the boot at all? Again,  a minor detail and perhaps the angle under which the video is taken?

The hairspring seems to run flat and concentric

Hmmm 🤔

 

Link to comment
Share on other sites

20 minutes ago, Endeavor said:

If in vertical positions a watch runs only between 140-180, that's IMHO too low for dynamic poising.

If you look at the numbers on the left, (heading "Fully Wound")it is running above 253° in all vertical positions. Which is good.

In dynamic poising, you then reduce the amplitude to about 140 - 180° to find the heavy spot. Those are the numbers on the right (heading "Amplitude 140-180°).   I reduced the amplitude to about 160°, and the numbers are the timings.

With the crown at the fastest position, the heavy spot is below the balance staff. 

Link to comment
Share on other sites

Unfortunately videos didn't open on my tablet. 

  Amplitude drop in vertical positions is expected cuz pivots ride on jewels ( hole jewels), therefore friction increase thus lower amp, but 160 is uncommonly low, so

Are numbers regarded as valid with 160 amp? 

 Side shakes algebrically add up in vertical positions, which can result in heavy pallete lock, affect guard pin' roller table clearance.

I would give another clean to jewels in escapement plus check side shakes therein. 

Rgds

 

  • Like 1
Link to comment
Share on other sites

The problem is NOT about amplitude drop in vertical positions - that is expected. The amplitudes for vertical positions  CU/CD/CL are 265,253,260° - all very good.

The 160° amplitude is BECAUSE I SET IT to determine the heavy spot on the balance.

The question is, why, when the horizontal rates are +5/+9s, are the vertical rates all so slow -32s, -46s, -89s ?

Link to comment
Share on other sites

...I know you can cripple a W10 if the spring for the center pinion isn't centered but this calibre isn't set up the same is it?

Is the escape wheel cap jewel arrangement running a bit dry in vertical positions? 

Edited by rehajm
Link to comment
Share on other sites

1 hour ago, mikepilk said:

If you look at the numbers on the left, (heading "Fully Wound")it is running above 253° in all vertical positions. Which is good.

Sorry, misunderstood 😉

 

Link to comment
Share on other sites

2 minutes ago, Endeavor said:

Sorry, misunderstood 😉

Thought you had. Dynamic balancing is interesting, but you can spend a lot of time getting a watch right.

The link you posted is my main reference point. Worth a read if you are interested.

What I really need are some small timing washers. Very hard to find. I've been looking on ebay for ages, but it's usually pocket watch washers that come up. 

  • Like 1
Link to comment
Share on other sites

 

1 hour ago, mikepilk said:

The link you posted is my main reference point. Worth a read if you are interested.

I’ve used that series of posts often. They take several passages from Jendritzki- a book I’ve found to be extremely valuable including so many diagnostic checks and fouls I haven’t seen mentioned anywhere. Several pages on regulators- pins, boots, the fancy ones, for example.

IMG_1529.thumb.jpeg.7be63d6bd1f2d4029264d42ec996cb0c.jpeg

Edited by rehajm
  • Like 1
Link to comment
Share on other sites

The hairspring is favoring the inside pin, but I don't think that's the issue. How does it look from the top with power off? From the vid there could be a chance it's just touching the stud at high amplitude, that would cause a significant gain, which would show up as a significant loss at lower amplitude; what's the rate horizontal at low amplitude? Sometimes there's a bit of spring sticking out the stud that can get bent toward the spring body and touch.

  • Like 1
Link to comment
Share on other sites

17 minutes ago, nickelsilver said:

How does it look from the top with power off?

You could have found it!  The hairspring was OK, but at some stage I must have touched the terminal coil and not noticed. It's hard to see under the cock, it's a big one (🤣),  but It's bunched up on the side of the stud - see stationary pic below. It looks like the coils could be touching each other at some point at high amplitude. When vertical they separate out more, hence the slower rate. 

That explains why the rates look so consistent at very low amplitude.

How did I not notice that 🥴

I feel foolish now, D'oh

WIN_20240311_15_53_43_Pro.thumb.jpg.938aee12ea57f5a618c04e385f5cc1ce.jpg

  • Like 2
Link to comment
Share on other sites

Haha- that’s a good catch- no look even…

…I have a Venus that I think is similarly touching somewhere once it gets to higher amplitudes. There’s a suspect burr on the collet next to the first coil but I can’t see if its hitting in situ…

Link to comment
Share on other sites

5 hours ago, Nucejoe said:

Unfortunately videos didn't open on my tablet. 

  Amplitude drop in vertical positions is expected cuz pivots ride on jewels ( hole jewels), therefore friction increase thus lower amp, but 160 is uncommonly low, so

Are numbers regarded as valid with 160 amp? 

 Side shakes algebrically add up in vertical positions, which can result in heavy pallete lock, affect guard pin' roller table clearance.

I would give another clean to jewels in escapement plus check side shakes therein. 

Rgds

 

The low amplitude is specifically chosen Joe. Both low amplitudes below 220° and high amplitudes above 220° will amplify any out of balance of the wheel, making it easier to find. There wlll come a point of too low an amplitude that makes the readings unreliable, too much portion of the swing taken up by the escapement. 

  • Like 1
Link to comment
Share on other sites

14 minutes ago, Neverenoughwatches said:

The low amplitude is specifically chosen Joe. Both low amplitudes below 220° and high amplitudes above 220° will amplify any out of balance of the wheel, making it easier to find. There wlll come a point of too low an amplitude that makes the readings unreliable, too much portion of the swing taken up by the escapement. 

More embaressing is ; that wasn't the only point of Mikes post I misunderstood, matey.

I think I best 😷.     and do more reading.

Ree Guards. 

 

 

Link to comment
Share on other sites

16 minutes ago, Nucejoe said:

More embaressing is ; that wasn't the only point of Mikes post I misunderstood, matey.

I think I best 😷.     and do more reading.

Ree Guards. 

 

 

Some extra reading never hurts, no need to isolate Joe but sometimes its nice to sit in a quiet chair away from all distractions for a few hours. Nev posted  a download of a good read a few days ago, maybe you picked it up. 

Link to comment
Share on other sites

42 minutes ago, Neverenoughwatches said:

Some extra reading never hurts, no need to isolate Joe but sometimes its nice to sit in a quiet chair away from all distractions for a few hours. Nev posted  a download of a good read a few days ago, maybe you picked it up. 

Did you just tell Joe to sit in the corner? 😅

Link to comment
Share on other sites

5 hours ago, mikepilk said:

You could have found it!  The hairspring was OK, but at some stage I must have touched the terminal coil and not noticed. It's hard to see under the cock, it's a big one (🤣),  but It's bunched up on the side of the stud - see stationary pic below. It looks like the coils could be touching each other at some point at high amplitude. When vertical they separate out more, hence the slower rate. 

That explains why the rates look so consistent at very low amplitude.

How did I not notice that 🥴

I feel foolish now, D'oh

WIN_20240311_15_53_43_Pro.thumb.jpg.938aee12ea57f5a618c04e385f5cc1ce.jpg

OK, but did Uou prove this? It is important to know for sure, not only to guess.

In the videos the hairspring seems not well centererd to the collet

Edited by nevenbekriev
  • Like 1
Link to comment
Share on other sites

9 hours ago, Endeavor said:

the hairspring breathing in the regulator pin/boot. It seems that the hairspring barely loosens itself from the regulator pin. It doesn't seem touching the boot at all?

As you say, hard to tell looking at the video. It needs to be assessed in 3D from various angles. Anyway, it reminded me of this post which I have found useful. Probably a bit too elementary for most of you. Nevertheless...

 

  • Like 1
Link to comment
Share on other sites

Not the case with your watch, but as for the hairspring "breathing", I do like the Etachron system a lot. One can turn the stud so that the hairspring runs exactly in the center/middle of the regulator, touching either side equally. One can reduce the gap of the regulator so that the duration at which the balance wheel slows down gets if even shorter.

Not having the Etachron, you have to manipulate the hairspring (and if required the distance between pin/boot) to achieve the same result / effect.

Glad you found the cause of your problem 😉

 

Edited by Endeavor
  • Like 1
Link to comment
Share on other sites

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.
Note: Your post will require moderator approval before it will be visible.

Guest
Reply to this topic...

×   Pasted as rich text.   Restore formatting

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.



  • Recently Browsing

    • No registered users viewing this page.
  • Topics

  • Posts

    • Hello and welcome fromLeeds, England, it looks like you are well on your way.  Multiple accounts are against forum rules, if you give the name on the other account it can be removed. 
    • That has become habit to quote since learning about electricity as a kid. I'm sure even my old school sparky (74) still uses divisions of 240v to work out current draw and fuse ratings. Uk  Rating on my travel kettle averages out at 230volts 😆
    • Hello everyone, My name is Matt and I like vintage watches along with pocket watches. I'm very new to the hobby of watchmaking, I've only been tinkering for about 2 months and I'm happy with some of the projects I've salvaged.  First I repaired This Waltham 17 jewel 18s and that was fun.  Then I repaired this old Zenith movement with 19 jewels. I couldn't find it on the pocket watch database though.  Then I repaired this one Lucerne 1 JEWEL movement. It was actually the first watch I got but it was so difficult to put back together with the fiscal amount of tools I had at the time so I just left it strewn in a Ziplock bag. Then My most recent project that came up a total win was this Longines Caliber 19AS, 17 jewels unadjusted. It turned out beautifully in my opinion and was by far the most difficult to reassemble.  I am in love with this hobby and its kind of engulfed my personality as of right now.  I plan on giving the Longines watch to my uncle, and so right now I'm trying to enjoy that watch the most I possibly can.    What I'm currently working on: A Seiko Automatic with a little date window and a weekday function too. I think its the 6309A.  Then there is this Omega pocket watch. Unfortunately I lost a little screw that held in the click spring. I'll have to source another one because its tiny and my room is filled with carpet, that thing is gone. Then theres this little tiny Elgin pocket watch. It's fully assembled but the amplitude is low and so I think it needs a new mainspring.  Then theres this Waltham 7 jewel movement that I also cleaned and reassembled, but it'll only run facing one way and again with the amplitude being low, I think it needs a new mainspring as well.  And finally, I'm working on this gorgeous Grade no.3 1883 Waltham Pocket Watch. It came with a broken balance staff on both sides and would only run when the watch was like sideways, crown to the sky. I'm really excited about that one though, might be worth something.      I've chosen to not show the other ones because they're not in movement holders right now, and the Omega is not put together.  As you can see, I take these pictures using the flash. Idk why but I like the way the flash makes the metal shine more than just my little lamp.  I'm also very eager to find a replacement for that balance wheel; any links would be appreciated.  From what I could tell on the Seiko, it's just dirty and It should run if I put it together carefully after cleaning.  Also, I fear that that 1883 Waltham may not run too good because of the mainspring, it's probably set because it's so old. I may seek a replacement for that too. I also need to find a crystal for that 18s gold filled Waltham that I've put together. It's just scary having it exposed.      I ALMOST FORGOT to mention. Is this Omega real?      You gotta turn your head a bit but it does indeed say "OMEGA", so I'm confused. I never knew they made pocket watches. And thats why I'm skeptical. Please let me know.    But thats pretty much it. I just wanted to put this out there so that I could begin using the website properly and also to let some people know I exist I guess.  Also this is my second account. I know that may violate some rules but I did the introduction wrong the first time and I wanted to fix it by making a new account. Please don't ban this one.    Thank you all, -Matt  
    • Sounds like you and @VWatchie were unlucky. Mine has been running on 230V UK for years  ( it's been 230V for many years @Neverenoughwatches, not 240V).  Though I have had to re-solder some wires. When I buy cheap Chinese electrical components I often take them apart to check they're not going to kill me. Since the demagnetiser, I bought a cheap fish tank heater for my plating set up. Plugged in, big flash and smell of burning. Deader. So I bought a cheap cup heater plate. Exactly the same - flash, bang, wallop.     
    • The old crystal is cracked on the edge, and I am looking for something like for like.
×
×
  • Create New...