timegrapher readings


Recommended Posts

Hi all ,

I recently bought my first timegrapher , i have watched many times marks videos about it and how to adjust.

My question is what means if i get various readings in one minute period and how many ms are "acceptable" in the timegraphers reading.

At my videos seiko 5 watch i get +6sd , +14sd ,and  -1 sd readings during a minute.i suspect that means i havent clean well or something ( pallet fork or escape wheell or something else) has a problem and have to start my detective job haha.

Since many of you have this machine long time i would appreciate some help.Thanks for your time :)

20161006_110341.mp4

Link to comment
Share on other sites

18 minutes ago, parmenion31 said:

At my videos seiko 5 watch i get +6sd , +14sd ,and  -1 sd readings during a minute.i suspect that means i havent clean well or something ( pallet fork or escape wheell or something else) has a problem and have to start my detective job haha.

Can you post few pictures? Videos are large to handle as attachment, better to upload to a dedicated platforum.

  • Like 1
Link to comment
Share on other sites

Looking at your vid the first thing i notice is your amplitude is low and the graph line is very erratic to start with have you de-magnatised the movement have you checked all pivots while cleaning making sure none are too worn or bent is the hairspring catching anywhere Low amplitude is usually something to do with the amount of power being transfered through the gear train and not kicking the balance wheel round with enough power seikos are usually have a lower amplitude but not that low I would check again all pivots and pivot holes are spotless.

  • Like 1
Link to comment
Share on other sites

Low amplitude on 7S26 is a common problem which is very difficult to correct. Happened to me quite a few times and cleaning things again and again did not helped of course. Neither replacing the main spring or barrel complete did help. I think it has more to do with 1. Having a perfectly shaped hairspring in order to deliver the maximum elastic impulse to the balance, and 2. The ideal angle and depth of the pallet stones. Both things are very difficult to get right for a beginner.

  • Like 1
Link to comment
Share on other sites

Also be careful holding your phone near the timegrapher as you will get noticeable interference.

 

I agree wit the others: get the amplitude up (was it fully wound?) and then start making judgments.

  • Like 1
Link to comment
Share on other sites

Here's a typical case I have got today, Seiko 6R15 which should be just 4 years old:

 

PA060084.JPG

Again, in my experience "cleanliness", magnetization, or interference have nothing to do with the problem. I will service this and I hope I can get it to normal, but sometime all one gets is a 20 deg. improvement, not enough when starting at 155.
In these cases unless the owner is OK with crappy and unpredictable timekeeping, all what can be done is swapping with for new movement, or try replacing escapement and balance.

 

 

Edited by jdm
  • Like 1
Link to comment
Share on other sites

Thank you all , you all helped , here some pictures promised to jdm.

Just now learning that amplitude is so serious, i was just focusing in hairspring adjustmunt and beat error.

I think no one answered how many ms are "acceptable" as an error.

For rodabob:   The watch needed a good cleaning , wasnt beating at all, was dead . I found nothing broken or marks of abusing.

20161006_215558.jpg

20161006_215631.jpg

20161006_215653.jpg

Link to comment
Share on other sites

9 minutes ago, parmenion31 said:

I think no one answered how many ms are "acceptable" as an error.

On current Seikos, approx 200 deg. is the very minimum. Swiss / better movements, 230. The more demanding the watchmaker, the higher the value.

NB: No need for you to struggle showing the actual movement on the machine.
I do that on selling pictures to show that is genuine :)

Edited by jdm
  • Like 1
Link to comment
Share on other sites

The M/s beat error is entirely up to you but with a better amplitude you should be able to adjust it nearer 0.0 but if your adjusting face down to 0.0 then face up will be different same as crown up and down will be different its all about finding that sweet spot in all positions no watch ever runs at 0.0 beat error in all positions you have to remember there's friction gravity and temperature to take into place too.

  • Like 1
Link to comment
Share on other sites



  • Recently Browsing

    • No registered users viewing this page.
  • Topics

  • Posts

    • attached is the tech /parts sheet. .   Can one ask how come it’s full of water. After being told that the waterproofing was not guaranteed.  Problem being electronics don’t mix especially withe the power source being attached.  You may be lucky but I should look out for a replacement movement or working donor to rob the movement from. C720.pdf
    • I thought this may be "right" forum to post this. After 7 months out in Switzerland I got my Navitimer back from Breitling. I had sent it off since its 20 years old to have a complete overhaul done... Nothing major wrong, but I did notice this seconds sundial skipping thingy.... $1300 later and 7 months later, I notice the 'skips' are still here unless the chronometer is engaged. I would think this is likely the hand binding up and maybe clearance with the dial, but at any rate Im interested as to whether this is something I should be returning the watch to Breitling for and insisting on being fixed. See below - 1 minute or so of regular running and then 1 minute with the chronograph engaged.  
    • A friend of mine just gave me an old pocket watch. The case is in very bad condition but the movement itself seems to be in a quite good shape. I can't identify which movement is this, since I have to find a missing winding stem. I attach few photos in a case somebody would be able to give some hints.  Thank's folks
    • Hello everyone, I recently received a request for advice on purchasing a vintage watch, a Waltham A-11 ORD. DEPT. I asked if they were aware that these types of watches have radioactive hands and dials... but they had no idea what I was talking about. So, I thought of creating a video that could help recognize them, starting from the distinctive details to using a Geiger counter. I hope you like it and find it useful for those venturing into the world of vintage watches. Bye.     