Murks,

The rate and amplitude look OK, and the amplitude should improve once the oils you have used get a chance to move bed-in, also I notice that you are using default 52 degrees for the lift angle, if you get the real lift angle (assuming it's not actually 52) this will change your amplitude - maybe higher, maybe lower.

I notice that the beat error is a little high, but not crazy high. At the risk of upsetting the purists, if the balance has an adjustment arm I would go ahead and try and get this <0.3 ms, but if it does not have an adjustable arm then I would probably leave well alone. Just my opinion.

  • Like 2
Link to comment
Share on other sites

Posted (edited)

Yea a have given it a full service this movement is chinese 2813 whats the lift angle for this movement does anyone knows  and how  do a remove the default  of 52 degrees on my timegrapher?

5 hours ago, Waggy said:

Murks,

The rate and amplitude look OK, and the amplitude should improve once the oils you have used get a chance to move bed-in, also I notice that you are using default 52 degrees for the lift angle, if you get the real lift angle (assuming it's not actually 52) this will change your amplitude - maybe higher, maybe lower.

I notice that the beat error is a little high, but not crazy high. At the risk of upsetting the purists, if the balance has an adjustment arm I would go ahead and try and get this <0.3 ms, but if it does not have an adjustable arm then I would probably leave well alone. Just my opinion.

 

5 hours ago, Waggy said:

Murks,

The rate and amplitude look OK, and the amplitude should improve once the oils you have used get a chance to move bed-in, also I notice that you are using default 52 degrees for the lift angle, if you get the real lift angle (assuming it's not actually 52) this will change your amplitude - maybe higher, maybe lower.

I notice that the beat error is a little high, but not crazy high. At the risk of upsetting the purists, if the balance has an adjustment arm I would go ahead and try and get this <0.3 ms, but if it does not have an adjustable arm then I would probably leave well alone. Just my opinion.

Hi you said my lift angle is default how do a remove this from my timegrapher?

Edited by Murks
Link to comment
Share on other sites



  • Recently Browsing

    • No registered users viewing this page.
  • Topics

  • Posts

    • Could it be an abbreviation for Denier? "Denier (/ˈdɛniər/) or den (abbreviated D), a unit of measure for the linear mass density of fibers, is the mass in grams per 9,000 metres of the fiber. The denier is based on a natural reference: a single strand of silk is approximately one denier; a 9,000-metre strand of silk weighs about one gram." The pivot gauge looks interesting, however I wouldn't trust myself to drive a pivot into a wedge incase it jams and gets damaged/snaps trying to get it free.
    • @clockboy The damage on the movement much worse. The images against the green background are my watch the other 2 are from the internet. The rotor has erased the perlage. I am surprised how is this watch still running.
    • Hi All, I recently bought a few gauges on Ebay. One of those was my actual target, a roller jewel gauge. Pictured below was one of the other gauges I received. From what I gather, this one is for measuring pivots. No sure if I'm right here. A confirmation of my assumption would be great. The gauge also has an abbreviation of "DEN". I have looked around and don't find this relative to watchmaking. Any clue to what this abbreviation means? Thanks, Don
    • Hello and welcome to the forum. Te attached file my help you repair a few. TZIllustratedGlossary.pdf
    • Thank you for your introduction and welcome to this friendly forum. We all look forward to your contributions and continued involvement. 
×
×
  • Create New...