Movement Serviced 2836-2 Runs perfect on Final Test Machine, but


Recommended Posts

Normally it is a positional error. Did you test the watch in all positions. Some wear a watch crown up some wear dial down some remove the watch at night etc etc. I have found the biggest culprit is the balance end shake is to great. If so and if you have the correct equipment you can adjust the lower jewel setting with fractions of a mm making a huge difference. Side shake is another head ache altogether. 

  • Like 2
Link to comment
Share on other sites

Yes CB, on the timegrapher, is all six positions, the movement is running and been regulated and is well within COSC, and its this which is confusing.  I will wear the watch for a few days once I resize the bracelet.  Its a head scratcher for sure.

Link to comment
Share on other sites

5 hours ago, SSTEEL said:

Checked everything, nothing amiss, only issues are when worn.  Its a very strange issue for sure.

You need to verify first in person. There is a say, no head bandage before breaking it.

Link to comment
Share on other sites

On 6/16/2018 at 9:20 AM, SSTEEL said:

Yes CB, on the timegrapher, is all six positions, the movement is running and been regulated and is well within COSC, and its this which is confusing.  I will wear the watch for a few days once I resize the bracelet.  Its a head scratcher for sure.

Check the end shake on the balance, very important. It must be perceptible to ensure that the balance is totally free. A touch of a good quality lubricant on the jewel holes and end stones is essential with regard to the balance.   Check the guard pin as well for correct length and operation. the Timegrapher should show any discrepancies up though. This sounds a little bit of a poser !!. I am probably teaching you to suck eggs !!.  Is the amplitude constant?  oldhippy gave me a tip not to oil the pallet arbors on an Everite watch and this improved the Bal amplitude from 140-190 to a steady 220deg !. Check the clearance on the pallet arbors  to be sure the pallet is quite Free. It`s rotational moment is so little that oil is not neccessary and was causing drag in my case. Hope this helps, Mike.

Link to comment
Share on other sites

19 minutes ago, ecodec said:

This sounds a little bit of a poser !!. I am probably teaching you to suck eggs !!

Mike, indeed SSTEEL is a professional watch repairer, check his previous postings. The good advice you gave applies when the timegrapher data is unsatisfying, but here we're discussing a different, and still unproven issue.

Link to comment
Share on other sites

There are two common problems that can cause this phenomenon of poor timekeeping.  What does the customer use to compare the watch to?  Specifically how does the customer know the watch is fast or slow. Sometimes customers will look at a variety of time sources none of which are on time and or use time sources that are less accurate than the watch.

Then this is an automatic the problem with automatics are they need movement to stay wound up. Not all customers are active enough to keep their watch wound up. As people get older they're not necessarily as active as they once were or the lifestyle changes variety of reasons they keep automatic watches from getting enough movement to stay wound up.

 

  • Like 2
Link to comment
Share on other sites

23 minutes ago, JohnR725 said:

There are two common problems that can cause this phenomenon of poor timekeeping.  What does the customer use to compare the watch to?  Specifically how does the customer know the watch is fast or slow. Sometimes customers will look at a variety of time sources none of which are on time and or use time sources that are less accurate than the watch.

Then this is an automatic the problem with automatics are they need movement to stay wound up. Not all customers are active enough to keep their watch wound up. As people get older they're not necessarily as active as they once were or the lifestyle changes variety of reasons they keep automatic watches from getting enough movement to stay wound up.

 

I agree to a certain extent but I am under the impression this is a large error. Watches run at different rates depending on their position but with a healthy watch the differences over 24hrs will be seconds not minutes. SSTEEL is going to have to wear the watch for a few days to see the true error.

Link to comment
Share on other sites

10 hours ago, jdm said:

Mike, indeed SSTEEL is a professional watch repairer, check his previous postings. The good advice you gave applies when the timegrapher data is unsatisfying, but here we're discussing a different, and still unproven issue.

Indeed jdm. 

Link to comment
Share on other sites

I will be wearing the watch over the next few days, so will report my findings.  I wear a Rolex Submariner as my daily watch, and it auto winds perfectly with my day to day activity working, and whatnot, so I will at least know if the issue is related to the owner not being active enough.  I have come across a few customers who work at a desk, and they are not active enough to auto wind their watch movements.  On more than one occasion, I have had to instruct the owner how to manually wind their watch movement, quite surprising, some customers were not even aware they could manually wind their watch. :blink:

Link to comment
Share on other sites

just a quick note to your last post SSTEEL, i have just serviced a job lot of watches for a family friend,one was her childhood watch, she's 48 and didn't know it was a manual/mechanical watch and i had to tell her how to wind it etc, couldn't believe I had to explain that point.

Link to comment
Share on other sites



  • Recently Browsing

    • No registered users viewing this page.
  • Topics

  • Posts

    • If frank says these bombè jewels are for balances and similar pivots then they are used cupped side facing the pivot and then capped i assume. In Nicklesilver's reply i guess that cup can also be reversed and used as an oil reservoir leaving the shoulder riding on a less frictioned domed surface. 🤷‍♂️ Hence the use of them on wheels closer to the escapement that have less torque requiring efforts to reduce friction and are also uncapped.
    • Looks that way, if it had any amount of significant jewel count it would say so on the dial.    Looks that way, if it had any amount of significant jewel count it would say so on the dial.  Same answer as Mike regarding the dial, its often just a slapped on name of a retailer but i thought you were way above this understanding Pete ?
    • Light naptha is what you need to be looking for. The carbon to hydrogen ratio is high with heavy naptha, it doesn't evaporate as well as light naptha and is prone to leaving more residue. Try tom's suggestion of the mirror test.
    • Its a big item Michael and a lot that could go wrong with the process of getting it to its new location. I once helped install one of these units into a conservatory that i had built. It was big and bloody heavy and took six of us to get it in safely. If you have sliding doors and you need to remove the whole frame then that may or may not go well. But i would advise getting a professional in if you have little knowledge on the installation of them. The glazing of the doors could well be 6mm laminated glass which again will be heavy very heavy if the glazing units are big. The glass would also need to be set up and packed correctly to open the doors properly more so if they are unframed and the sliders are upvc. Seems a shame but if no one is biting and that may be due to having no warranty i guess the only takers might be someone that installs them for a living and knows how to repair. If its fibreglass just be careful, i sometimes have to cut fibreglass baths out and they are nasty to work with . Thick gloves are definitely recommended,  not fun picking out fibreglass spells two days later that have started to go septic. Good luck with it all matey.
    • I don’t know anything about this product, however assuming you have it if you put a drop on a mirror or piece of glass lying flat on a table and leave it to evaporate if it leaves any sort of residue it is not really suitable for watch part cleaning.   Tom
×
×