7s26 ticking too quiet for timegrapher ?


dreja

Recommended Posts

Hello,

I'm wondering what is your experience with a 7s26 watch, when you put it on the timegrapher ? Does the ticking sound have enough volume, to be picked up by the microphone ?

I'm using a standard Chinese timegrapher.

When I put the whole watch on the microphone (for eg. my SKX007 or my colleagues seiko 5), the microphone has difficulties to pick up the ticking sound. If I put on only the movement without the case, there are no problems.

 

So I'm wondering where could be the problem:

The spacer between the case and the movement is plastic, which "isolates or weakens the sound flow" to the case ?

As the crown is located at 4 and the case is not symetric. The problem might occur because the edge of the case is not symetrically pressed to the microphone ?

 

I don't have issues with other "low volume ticking" watches or with bare movements without the case. So I hope my microphone works ok.

Thanks

Andrej

Link to comment
Share on other sites

10 hours ago, dreja said:

When I put the whole watch on the microphone (for eg. my SKX007 or my colleagues seiko 5), the microphone has difficulties to pick up the ticking sound.

That is divers' watch with a thick case and caseback, so maybe that is the reason.

However I have no problem at all getting a good signal on a closed 7S26-0020 with either a 1000 or 1900 model timegrapher.

Link to comment
Share on other sites

Hello,

thanks for the answers.

I've got a Timegrapher No. 1000, about 8 years old. It seems, that there are at least 2 versions of the No. 1000 model. The difference can be seen, when you switch on the display:

1st version shows the daily rate only on the top of the screen.

2nd version shows the daily rate also on the main screen in 4 different horizontal display sections.

 

I have also noticed that there are 2 types of microphone stands. One stands is more rectangular, while the other is more circular.

If you look at the back, there are two versions of Frequency calibration connector. One has a female BNC connector while the other has a usb square connector.

 

I've got the first version with a BNC connector and a rectangular microphone stand.

Unfortunatly my version has not the ability to adjust the microphone gain. The up/down buttons change only the contrast (both in capture mode and while stopped)...

 

Andrej

 

Link to comment
Share on other sites

I agree it's probably just the thickness of the case although there may be a method to adjust the sensitivity.  I've done some Seiko 7002's that had pretty thick cases and found that taking the case back off during testing was an acceptable workaround.

Link to comment
Share on other sites

8 hours ago, Pip said:

I think you can only adjust the mic gain on the 1900 version.

Yes, and on the 1000 model also as mentioned above.

However, using either model I have cheched literally hundreds of 7S26 watches with all types of cases and never had to adjust gain in order to pick a good signal.

Link to comment
Share on other sites

1 hour ago, jdm said:

Yes, and on the 1000 model also as mentioned above.

However, using either model I have cheched literally hundreds of 7S26 watches with all types of cases and never had to adjust gain in order to pick a good signal.

Oh, I'm probably wrong then as I thought you couldn't adjust gain on the 1000.

As an aside I've not had issued picking up any ticking and tocking from diver cased watches either so maybe it's a mic problem.

Link to comment
Share on other sites

Thanks for all answers, I have found the issue :)

The problem was, that I have put some Tesa adhesive tape on the one microphone's plastic part. I did this to avoid scratching the watch - without the protection tape I could see (with loupe) the small marks on watch case.

Without the protection tape it works ok :) Instead of the tape I can put a piece of paper or even a paper sticker (as on the photo) between the watch and microphone.

 

 

IMG_20180616_084902.jpg

Link to comment
Share on other sites

15 minutes ago, dreja said:

 I did this to avoid scratching the watch - without the protection tape I could see (with loupe) the small marks on watch case.

Very good, you're treating a $200 watch same as $2,000 one. Which can be justified by the OCD attitude of some owners.

Link to comment
Share on other sites

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.
Note: Your post will require moderator approval before it will be visible.

Guest
Reply to this topic...

×   Pasted as rich text.   Restore formatting

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.



  • Recently Browsing

    • No registered users viewing this page.
  • Topics

  • Posts

    • OK, welcome in the world of alarm clocks... I guess the 4th wheel is dished because it is from another movement. If it was not dishet, then it would not mesh with the pinion of the escape wheel, am I right? The marks of wear on the 4th wheel pinion doesn't corespond to the 3th wheel table position, at list this is what i see on the picts. Calculating the rate is easy - there is a formula - BR = T2 x T3 x T4 x T5 x 2 /(P3 x P4 x P5) where T2 - T5 are the counts of the teeth of the wheels tables, and P3 - P5 are the counts of the pinion leaves. Vibrating the balance is easy - grasp for the hairspring where it should stay in the regulator with tweasers, let the balance hang on the hairspring while the downside staff tip rests on glass surface. Then make the balance oscillate and use timer to measure the time for let say 50 oscillations, or count the oscillations for let say 30 seconds. You must do the free oscillations test to check the balance staff tips and the cone cup bearings for wear. This kind of staffs wear and need resharpening to restore the normal function of the balance.
    • Glue a nut to the barrel lid, insert a bolt, pull, disolve the glue.  Maybe someone will have a better answer. 
    • The stress is the force (on the spring) x distance. The maximum stress is at the bottom, and decreases up the arm. That's why they always break at the bottom. I used a round file, then something like 2000 grit to finish. I gave the rest of the arm a quick polish - no need for a perfect finish. Just make sure there are no 'notches' left from cutting/filing. The notches act like the perforations in your toilet paper 🤣
    • It's probably a cardinal rule for watch repair to never get distracted while at the bench. Yesterday, after finishing a tricky mainspring winding/barrel insertion (I didn't have a winder and arbor that fit very well) I mentally shifted down a gear once that hurdle was passed. There were other things going on in the room as I put the barrel and cover into the barrel closer and pressed to get that satisfying snap. But when I took it out I realized I never placed the arbor.  When opening a barrel, we are relying on the arbor to transfer a concentrically-distributed force right where it is needed at the internal center of the lid. However, when that isn't present it's difficult to apply pressure or get leverage considering the recessed position of the lid, the small holes in the barrel and the presence of the mainspring coils. It was a beat-up practice movement so I didn't take a lot of time to think it over and I pushed it out using a short right-angle dental probe placed in from the bottom, but that did leave a bit of a scratch and crease in the thin lid. I had also thought about pulling it using a course-threaded screw with a minor thread diameter smaller than the lid hole and a major diameter larger, but that may have done some damage as well.  Thinking about how this might have been handled had it been a more valuable movement, is there a method using watchmaking or other tools that should extract the lid with the least damage? 