13 minutes ago, CWRNH said:

What is the common?

It is my empirical and partly technical opinion working with sound. But above 40 dB it is very good. Between 45 and 50 is almost perfect. More than 50 would be the ideal.
The WOS screen is also not used as a reference, as it shows very little noise and the peaks are very high. Furthermore, WOS does not apply any gain. What I interpret that determines that the signal is very good.

  • Like 1
Link to comment
Share on other sites

1 minute ago, guidovelasquez said:

It is my empirical and partly technical opinion working with sound. But above 40 dB it is very good. Between 45 and 50 is almost perfect. More than 50 would be the ideal.
The WOS screen is also not used as a reference, as it shows very little noise and the peaks are very high. Furthermore, WOS does not apply any gain. What I interpret that determines that the signal is very good.

So far the best I have had is 55

Link to comment
Share on other sites

2 minutes ago, guidovelasquez said:

It seems to me an excellent result. In my case I do not exceed 52 dB. But that is enough.

Here is WOS with a Val7750 movement, SNR is not as good, it is at 30.  Probably because the movement is so quiet and inside its heavy case.image.thumb.png.79c0e456d6edba2f3ae3ad859a439f65.png

Link to comment
Share on other sites

2 minutes ago, guidovelasquez said:

I suggest you do a test with eTimer, disabling the filter. With this you can see what is the real value of the noise signal coming from the mic + pream.

Was your 52 results with or without filtering?

Link to comment
Share on other sites

18 hours ago, CWRNH said:

What Spectrum analyzer are you using?

I used audacity for those pictures.  But I have also added ability to generate spectrograms into tg-timer, https://github.com/xyzzy42/tg/tree/trentpi/python

Looking at sound again, it seem very clear that something is ringing inside the movement, but this happens only on every other beat.  Maybe pallet fork is vibrating after each beat, but only in one direction, due to some difference in the pallet jewels?  Or escape wheel vibrates depending on which pallet jewel is locking it?

image.thumb.png.fd6b82b9b31af3a8231df9cc4abae491.png

image.thumb.png.af99a751b2570efa0887e3b17c826e71.png

Edited by xyzzy
  • Like 2
Link to comment
Share on other sites

19 hours ago, xyzzy said:

I used audacity for those pictures.  But I have also added ability to generate spectrograms into tg-timer, https://github.com/xyzzy42/tg/tree/trentpi/python

Looking at sound again, it seem very clear that something is ringing inside the movement, but this happens only on every other beat.  Maybe pallet fork is vibrating after each beat, but only in one direction, due to some difference in the pallet jewels?  Or escape wheel vibrates depending on which pallet jewel is locking it?

image.thumb.png.fd6b82b9b31af3a8231df9cc4abae491.png

image.thumb.png.af99a751b2570efa0887e3b17c826e71.png

I downloaded the source code...  No idea how to use it!  LOL  Any guidance would be greatly appreciated...

  • Like 1
Link to comment
Share on other sites

3 hours ago, CWRNH said:

No idea how

Well, if you are interested in investing a few hours here goes:
Source code is like the electronic diagrams of an electronics project. You must run them. But you can modify it according to your goals. Thus, from the source code, you will produce your own version of the program or, if you do not modify anything in the source code, produce the program in someone else's version.
Necistaras as in my case.
1.- Atom to rewrite the source code.
2.- MSYS2 to "compile" the program.
The guidance on how to proceed is in this link

https://github.com/vacaboja/tg

and here I put a clipping of where you should look.

github.JPG.a097ec94ee98072d8822096961d781a5.JPG

  • Like 1
Link to comment
Share on other sites

4 hours ago, guidovelasquez said:

Well, if you are interested in investing a few hours here goes:
Source code is like the electronic diagrams of an electronics project. You must run them. But you can modify it according to your goals. Thus, from the source code, you will produce your own version of the program or, if you do not modify anything in the source code, produce the program in someone else's version.
Necistaras as in my case.
1.- Atom to rewrite the source code.
2.- MSYS2 to "compile" the program.
The guidance on how to proceed is in this link

https://github.com/vacaboja/tg

and here I put a clipping of where you should look.

github.JPG.a097ec94ee98072d8822096961d781a5.JPG

thank you

Link to comment
Share on other sites

On 4/30/2021 at 3:15 AM, CWRNH said:

image.thumb.png.d695fc8a685a85144e10173fee5ea9b4.png

What amp design have you used to achieve this result?

Also, is this still using the clip-on piezo mic?

I've tried to look back at this huge thread, but not really got the full information.

Thanks!

 

Link to comment
Share on other sites

1 hour ago, Bonzer said:

What amp design have you used to achieve this result?

Also, is this still using the clip-on piezo mic?

I've tried to look back at this huge thread, but not really got the full information.

have you look at the website for the software he has the official amp.

then just because a lot of us had our own designs sprinkled through the current 44 pages of discussion. Which probably does make for interesting and confusing reading

then there's the confusion factor for all of us? The picture above looks nice doesn't it but were missing details as you've asked about. Like a pocket watch will look nicer than a wristwatch. We really all need to buy the exact same watch for running tests so we all of the same thing and we can compare. or we need to convince the Chinese to clone the witschi test module so we all have the exact same thing so pictures like the one posted above would have meaning because we could all try the same thing and see that ours does the same thing.

Link to comment
Share on other sites

On 5/1/2021 at 8:00 AM, CWRNH said:

I downloaded the source code...  No idea how to use it!  LOL  Any guidance would be greatly appreciated...

If you go to the page and scroll down, there are instructions for building for a number of different operating systems.  I have written more instructions for Linux and MacOS, but do not use Windows, so I can not improve those myself, which is one of the reasons they are probably the worst.  Also because Windows makes these things harder.

I can and do provide installable packages for Linux, where to get those is also described if one scrolls down, but do not know how to do this for Windows or have a Windows computer to do it on.

Link to comment
Share on other sites

17 minutes ago, JohnR725 said:

We really all need to buy the exact same watch for running tests so we all of the same thing and we can compare.

We do not necessarily have to have the same clock, the same amplifier or sensor or microphone.
Reasonable results can be obtained in a number of ways. Also keep in mind that programs or software can handle very poor sounds. Of course the ideal is a good sign.
Here I leave the results and the photos as I obtain them.

1960383186_WhatsAppImage2021-05-04at13_10_37.thumb.jpeg.9a5d22666570488bbd0a121cd2d1108c.jpeg

546313470_WhatsAppImage2021-05-04at13_10.37(1).thumb.jpeg.c2328a1d3dff966492084f26f87fb5ac.jpeg

1292676605_APOLOGA.PNG.1a5fb29796d307f9902ef88585b5f163.PNG

rec0504-131406.wav

 

  • Like 1
Link to comment
Share on other sites

20 hours ago, guidovelasquez said:

We do not necessarily have to have the same clock, the same amplifier or sensor or microphone.
Reasonable results can be obtained in a number of ways. Also keep in mind that programs or software can handle very poor sounds. Of course the ideal is a good sign.
Here I leave the results and the photos as I obtain them.

personally I still like the witschi test device but it probably cost an arm and a leg and six fingers and your firstborn.

so I'm intrigued and enthusiastic about your idea but were still stuck with what I'm trying to get at is we need a reference device. It should simulate a pocket watch and a little tiny ladies watch with set parameters of all of us in the entire group put it on our what ever we all would get the exact same results which is hard to do if everyone has everything different.

The problem I've seen when answering questions on the discussion group at least watch repair questions is my watch is doing this well is it really? Are we basing answers on faulty diagnostics. This actually happened in a rather lengthy discussion with a little tiny ladies watch where we just weren't sure where we stood with us faulty diagnostics or faulty watch. We needed a tiny reference to verify that the timing whatever was working.

Or here when people put their circuits together how do they know it's actually working out of they know it's sensitive enough.

I was just thinking of you all have the same watch at least we would all have the same results but you indicate that's not necessary so I'm looking forward to your ideas of solving this problem.

then probably too many variables but I wonder if we could take your wave file for instance play it on a computer perhaps a small speaker because it has a moving coil I wonder if that could simulate a watch? But were still stuck with a standard if your computer is outputting the volume at twice the level of mine were measurement results may look different.

 

  • Like 1
Link to comment
Share on other sites

6 hours ago, JohnR725 said:

We needed a tiny reference to verify that the timing whatever was working.

Well, you're right. We would need a very reliable standard to do our tests.
Here I have a proposal. It is not created by me. It comes with Graham Baxter's eTimer software. It is a waveform in wav format.  Is 2min and 40 seg, the sound of a 18000 BPH or 5BPS clock. Executed properly and with any of our timing machine programs, it should give the following results. -1 second of disface per day. 0.0 beat error and 262 degrees of amplitude when the lift angle is set at 52 degrees. Not much but it has helped me to learn.

1574735323_3Programas.thumb.JPG.f16bcd677d77afdcd7d0a2a9936e2638.JPG

 

Edited by guidovelasquez
delete sound file.
Link to comment
Share on other sites

9 hours ago, JohnR725 said:

we need a reference device

Frankly, the reference device is an atomic clock.  So most, if not all, of this thread is OCD.  Interesting...fun...and useful in a vacuum, but at the end of the day, a mechanical movement pales in comparison to digital electronics. So...why bother? 

Fine, I am OK with doing your best, but polishing a turd has no real benefit.  Accept that mechanical watches are not accurate and move on.  Value the beauty of their architecture and reset them once per week.  Rent or download the movie "Castaway" and face reality!

Link to comment
Share on other sites

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.
Note: Your post will require moderator approval before it will be visible.

Guest
Reply to this topic...

×   Pasted as rich text.   Restore formatting

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.




  • Similar Content

  • Recently Browsing

    • No registered users viewing this page.
  • Topics

  • Posts

    • It's probably a cardinal rule for watch repair to never get distracted while at the bench. Yesterday, after finishing a tricky mainspring winding/barrel insertion (I didn't have a winder and arbor that fit very well) I mentally shifted down a gear once that hurdle was passed. There were other things going on in the room as I put the barrel and cover into the barrel closer and pressed to get that satisfying snap. But when I took it out I realized I never placed the arbor.  When opening a barrel, we are relying on the arbor to transfer a concentrically-distributed force right where it is needed at the internal center of the lid. However, when that isn't present it's difficult to apply pressure or get leverage considering the recessed position of the lid, the small holes in the barrel and the presence of the mainspring coils. It was a beat-up practice movement so I didn't take a lot of time to think it over and I pushed it out using a short right-angle dental probe placed in from the bottom, but that did leave a bit of a scratch and crease in the thin lid. I had also thought about pulling it using a course-threaded screw with a minor thread diameter smaller than the lid hole and a major diameter larger, but that may have done some damage as well.  Thinking about how this might have been handled had it been a more valuable movement, is there a method using watchmaking or other tools that should extract the lid with the least damage? 
    • 🤔 what happens if lubrication is placed directly on top of epilame ? Making a small groove so the lubrication doesn't spread across the component but what if when lubing a little overspills and sits on the epilame .
    • Why just the bottom mike ? Is it worth polishing the whole arm ?
    • The one thing I took time over was to round and polish the curves at the bottom of the jumper arm. The slightest mark (left over from cutting of filing) acts as a stress raiser, just where you don't need it. 
    • I printed the base and it is a bit too large to fit on the base of the hand setter. The ring bumps into the column so it is unable to full seat on the central ring. I'm going to try removing some of the materal and see if I can't get it to go down.
×
×