Oh no, can't use my Bergeon micrometre! Help needed!


VWatchie

Recommended Posts

image.png.e25961afb8df7d27c9d228f960ca476f.png

image.thumb.png.c1e5de6a24e89c5d7e307e45dca3de78.png

image.thumb.png.8a1b9a2a8076a39210efac4e7bc7d40a.png

image.png.9da7261626780a97c34d438661f200fb.png

image.png.570aeb4bf77de5ba2864b12ad968c640.png

Some time ago I bought a Bergeon micrometre (and a Seitz jewelling tool) to be able to replace damaged jewels. The micrometre works perfectly down to the exact hundredth of a millimetre but trying to measure my first jewel I bumped into a very unexpected problem; the height adjustable round table fouls the spindle (circled in red).

I don’t think, or rather I hope, that the micrometre isn’t out of order and can be adjusted so that the spindle does not bump into the table. But searching high and low I just can’t find any documentation or user manual for this type of micrometre. Being a hobbyist having spent a small fortune on these tools, this situation is very frustrating.

So, the question is, what can I do to adjust my micrometre? As I have no clue what the various parts of the micrometre are named, I have given the various parts in the pictures a letter (in yellow) so that if anyone who knows how do this (and takes pity on me :unsure:) more easily can guide me.

Keeping my fingers crossed. Hard!

image.png

Edited by VWatchie
Tried to remove the last picture at the bottom of the post
Link to comment
Share on other sites

Hi Watchie   from what I can see  and not ever having used one of these, the table is for supporting parts being measured and in no other way interferes with the actual measurement. So by fitting a shim or disc with a lesser diameter to the table top to allow the arm to clear the table and support the piece being measured  may solve the problem . 

Link to comment
Share on other sites

it looks to me that if you loosen the two screws A & B then move the micrometer away from the table, move shaft C, which is now loose to the middle of the table E.

Tighten the two screws A & B 

You should the have the meeting point of the two shafts in the middle of the table and both point will be able to meet each other.

Edited by Alaskamick
modify
  • Like 4
Link to comment
Share on other sites

16 minutes ago, watchweasol said:

Hi Watchie   from what I can see  and not ever having used one of these, the table is for supporting parts being measured and in no other way interferes with the actual measurement. So by fitting a shim or disc with a lesser diameter to the table top to allow the arm to clear the table and support the piece being measured  may solve the problem . 

Thanks for the suggestion and the thought had occurred to me (lo and behold ;)) but I really don't think that's how it was designed and meant to be used. @Mark uses a similar micrometer, and here you can see how the table is supposed to support the jewel while being measured. I really don't want to add an extra disc/shim as it will introduce another piece to handle, and I'm not even sure how well it would work. Well, as a last resort I guess it could be OK.

Edited by VWatchie
Link to comment
Share on other sites

11 minutes ago, Alaskamick said:

it looks to me that if you loosen the two screws A & B then move the micrometer away from the table, move shaft C to the middle of the table E. you   should the have the meeting point of the two shafts in the middle of the table and both point will be able to meet each other.

Yes, I've considered that as an option too, but then the scale wouldn't be correct. I guess if I move it a millimeter I could always subtract a millimeter from the reading, but sooner or later I'm going to forget and mess it up. Anyway, thanks for the suggestion and I might try it if a better (or more correct) way doesn't come up.

Link to comment
Share on other sites

You shouldn't have any trouble. Once you have the two spindle flats touching in the centre of the table all you have to do is adjust the location of the micrometer scale, as with any micrometer they can be altered easily. It is normal practice.

 

I didn't notice that the ends of the two spindles were Flat.

Just bring them together after loosening the two screws earlier mentioned, then you adjust the scale by loosening collar H, line the micrometer up where you can see it easily then retighten H.

Screw L may lock the micrometer to stop it turning so that will need ti be loosened a little

 

Get the LH side in the right place first with the flat in line with the centre of the table then move to the micrometer side and start altering that.

Edited by Alaskamick
Link to comment
Share on other sites

if i am correct you should be able to adjust the left pin or what to call it . By unscrewing  the A and B screws . never mind the right pin isn't zero . That you can adjust later.  The big screw on the ratchet is only for  taking the click mechanism  apart inside the handle .  To adjust the zero on the scale loosen the big hex nut on the  back of the scale . And turn the scale to zero .  That way you will get clearance  for the table. 

  • Like 1
  • Thanks 1
Link to comment
Share on other sites

41 minutes ago, Alaskamick said:

it looks to me that if you loosen the two screws A & B then move the micrometer away from the table, move shaft C, which is now loose to the middle of the table E.

Tighten the two screws A & B 

You should the have the meeting point of the two shafts in the middle of the table and both point will be able to meet each other.

I believe after he moves C, he'll have to adjust H,L and K to re-establish his zero point, which is C and F touching face to face. EDIT: @rogart63 has it correct!

Edited by FLwatchguy73
corrected info
  • Like 1
Link to comment
Share on other sites

2 minutes ago, FLwatchguy73 said:

I believe after he moves C, he'll have to adjust H,L and K to re-establish his zero point, which is C and F touching face to face. EDIT: @rogart63 has it correct!

Maybe not 100% . The K can be used to align the zero to the scale to zero .  It no harm in trying . As long as you  arne't loosening tha screws to much.  Look at a  normal micrometer . They work the same way . 

  • Thanks 1
Link to comment
Share on other sites

You can definitely move the fixed anvil toward the mic head (to the right). Just enough that you clear the table. On one of mine there is only  maybe 0.10mm clearance, but that's fine. Once you've done that show us what it's doing on the scale and we can advise.

 

The collar H is to tighten a collet that takes up the play on the spindle. You won't need to touch that.

 

 

  • Thanks 1
Link to comment
Share on other sites

image.thumb.png.9490c5b72ff2d70760b53c35141321de.png

image.thumb.png.7b08bfcf78e130e5789865e90193be84.png

image.png.eda5ca316c2eaf82313ff1cec55c9cbe.png

Thanks guys! :thumbsu: All your support really makes me warm at heart! :wub:

As you can see, there has been some progress. (being someone who barely held a screwdriver in hands before the age of 55 I feel pretty proud of myself). Anyway, can't get the scale right. I loosened the big hex nut (even removed it as you can see) but the nut seems to be fixed to the screw, and no matter how I fiddle with it nothing happens. I guess something else has to be done to get it right...

Keeping my fingers crossed you can help me again!

Link to comment
Share on other sites

Put the screw/nut back on, you don't need to move the thimble (the barrel with the graduations 0-99). For future reference, it's held in place with that screw/nut, but is fitted on a taper and needs some coaxing to get off and move around.

 

If you loosen screw K you can move the inner sleeve, it will just slide around. Get it where you want it and tighten K and you're done.

  • Thanks 1
Link to comment
Share on other sites

The nut you have undone is only part of the micrometer mechanism, nothing to do with the mounting of the unit.

As I said in my earlier post try loosening screw L which I think locks it in place. also try undoing the  Ring H.

Something frees it up and allows it to move back and forth

Edited by Alaskamick
Link to comment
Share on other sites

image.thumb.png.a3d82b348f2a2569683e21a6c070c616.png
Mission accomplished. The Eagle has landed!

image.png.b9dd3e3da90e2ecf1b0cd9e7e62bec0b.png

So, I’m all good to go! Just can’t thank you enough guys! I know that for you seasoned repairers this is nothing special, but me, well, I feel like I’ve successfully taken off in a rocket! :lol:

For those of you finding this thread in the future here’s some additional info that I learned on the way.

After having removed screw A and B the spindle C just wouldn’t move pulling it by hand. So, I tapped on it ever so lightly horizontally in the direction towards spindle F and it moved in increments. To tap I used the plastic part of the hammer in the picture.

I also needed to rotate the spindles C and F slightly. I did this by grabbing on to the thick part of the left spindle C (after having tapped it out a bit) with pliers with a bit of transparent and though plastic film (zip bag) in between the pliers and the spindle. To rotate the right spindle F I simply used a small BAHCO adjustable wrench.

Loosening screw K as @nickelsilver and  @rogart63 point out is indeed the way to get the scale back to normal! The inner sleeve can then slide as well as rotate without affecting the thimble (the barrel with the graduations 0-99).

I’m really thrilled to own this beautiful precision tool! It gives me the same feeling that I get from looking at a really fine watch movement.

So again, a really big thank you to all of you! :thumbsu:
 

  • Like 5
Link to comment
Share on other sites

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.
Note: Your post will require moderator approval before it will be visible.

Guest
Reply to this topic...

×   Pasted as rich text.   Restore formatting

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.



  • Similar Content

  • Recently Browsing

    • No registered users viewing this page.
  • Topics

  • Posts

    • Thank you all for the replies!  Very informative! True enough, the Gamsol took some time to evaporate and does leave a residue. So not all naphtha are created equal!  Need to find alternatives then. i was able to try Hexane recommended by Alex and it seems great.  I wonder what the cons are?
    • Yeah, I saw that in the tech sheet but I don't see how it can be adequately cleaned with the friction pinion still in place. I've accidentally pulled the arbor right out of the wheel once when I used a presto tool to try and remove it. Mark shows how he does it with the Platax tool. Those are a little too pricey for me so I got one of these from Aliexpress and I just push down on the arbor with the end of my brass tweezers. That usually gets it most of the way out and then I just grab the wheel with one hand the and the friction pinion with the other and gently rotate them until it pops off. Probably not the best way but it's seemed to work for me so far.    
    • Thanks, Jon Sounds like a plan. Obviously I'll have the face on so do you think gripping with the holder will create any problems, but I will check in the morning to see how feasible it is but I assume it only needs to be lightly held. As for holding the movement instead of the holder won't be possible in this scenario as one hand will be puling on the stem while the other pushes the spring down. That was my initial concern is how the hell can I do this with only one pair of hands. All the other times I've had to remove the stem hasn't been a problem, apart from the force required to release the stem from the setting lever, but now I need to fit the face and hands its sent me into panic mode. If it had the screw type release things would be a lot simpler but that's life 😀   Another thing I will need to consider is once the dial and hands are fitted and the movement is sitting in the case I will need to turn it over to put the case screws in. I saw a vid on Wristwatch revival where he lightly fitted the crystal and bezel so he could turn it over, is this the only option or is there another method?      
    • Hi Jon, do You think that relation spring torque - amplitude is linear? I would rather guess that the amplitude should be proportional to the square of the torque. I had once idea to check it, but still haven't.
    • I did not. I thought about it, but I had cleaned it in my ultrasonic, and the tech sheet shows lubricating it in place already assembled, so I figured discretion was the better part of valor. Although since I have to depth the jewels anyway, maybe I pull the pinion off to rule it out 100% as part of the problem. Do you know if there's a safe way to do it? I don't want to use a puller because it would push down on the plane of the wheel, and that seems like a Bad Idea. I thought about using a roller table remover, but I don't think I have a hole stake pointy enough to push it down.
×
×