Unstable timegraph readings


Lc130

Recommended Posts

Hi All

I'm a beginner.  I ruined a hairspring on an otherwise fine bifora 115/1.  I replaced the balance, cock and all, from a donor (I broke off the curb pin in the original) .  Before the swap, the watch was stable on the time grapher.  Now I get varying readings.  I did demagnetize.

rate. amp, beat error

DU 

74, 250, 1.4

95, 228, 1.4

82, 249, 1.4

DD

70, 235,1.6

80, 250, 1.4

77 259, 1.4

PU 

100, 171, 3.9 

134, 180, 3.7

94, 205, 3.5

PD

-154, 197, 0.7

-73, 192, 0.7

-41, 198, 0.6

Any idea how I should proceed?  

Thank you

Charlie

 

 

 

 

Edited by Lc130
Link to comment
Share on other sites

Hi  The fact that the origional was ok and the only thing changed was the balance complete with the cock would point to some discrepency in the donor balance. There are two ways of solving this  1 strip the donor balance and use the parts to rebuild the origional or    2   the purchase of a balance complete from the supplier of your choice.d

Link to comment
Share on other sites

16 minutes ago, watchweasol said:

Hi  The fact that the origional was ok and the only thing changed was the balance complete with the cock would point to some discrepency in the donor balance. There are two ways of solving this  1 strip the donor balance and use the parts to rebuild the origional or    2   the purchase of a balance complete from the supplier of your choice.d

Aren,t hairsprings, balance wheel interchangable?

Link to comment
Share on other sites

Out of round balance wheel would cause variation in PU & PD positions. Poise issue is of the same nature.

If balance wheels weigh the same, the good HS + good balance wheel would work (simply interchangable).  In case of  different weight, there is still a chance to vibrate the two.

Have you checked the upper jewel assembly? Combine with excess endshake, they exacrebate the problem.

Best

 

 

  • Like 1
Link to comment
Share on other sites

11 minutes ago, watchweasol said:

Just a thought, did the donor watch work before you used the balance ?. If so put the balance back in the donor and checki it out on the vibrograph  it may have exibited this fault in the donor and you have transfered it to your watch.

Suppose it dose exhibit the same fault, how would this new info help  diagnos the fault or fix it. 

Link to comment
Share on other sites

It wouldnt diagnose the fault but it would prove that the balance assy was exibiting the fault before fitting to the recipient watch therefore a fault was transferred and was inherent in the donor watch before removal and refitting, There is no way of knowing if this was the correct balance that was in the donor in the first place.

  • Thanks 1
Link to comment
Share on other sites

3 hours ago, watchweasol said:

It wouldnt diagnose the fault but it would prove that the balance assy was exibiting the fault before fitting to the recipient watch therefore a fault was transferred and was inherent in the donor watch before removal and refitting, There is no way of knowing if this was the correct balance that was in the donor in the first place.

Thanks for the explanation. I see now, it saves us a search in the wrong place. 

Best 

 

Link to comment
Share on other sites

7 hours ago, watchweasol said:

Just a thought, did the donor watch work before you used the balance ?. If so put the balance back in the donor and checki it out on the vibrograph  it may have exibited this fault in the donor and you have transfered it to your watch.

Thank you.  It worked but I only tested the dial up position on the timegrapher before transferring it.  Next week I plan move it back to the donor and test all positions.  

Charlie

 

Link to comment
Share on other sites

Hi LC  Thats good. Once thats done and proved either good or faulty we are then in the situation where we have established facts and can move on. Problem with donor watches we have no history, some may have been cobbled together to sell regardless.  In terms a working watch is defined by the fact that the balance swings and it ticks, There may be a multitude of sins in the watch yet to be discovered.  when buying donor watches it is a case of buyer beware (caveat emptor), but we all do it knowing  most of the time it works the odd time it doesnt.

Link to comment
Share on other sites



  • Recently Browsing

    • No registered users viewing this page.
  • Topics

  • Posts

    • Hello and welcome from Leeds. 
    • Unfortunately I'm not that lucky. I started on the train side and after I noticed the binding I pulled everything out except the driving wheel to rule everything else out. It still binds. I'm going to double check that the pinion is fully seated on the staff first, then if no joy I'll push the bridge jewel up a fraction of a mm. Fingers crossed!
    • Happy to have helped, great way to start the day with a win! 🥳
    • Thank you for the advise!! It worked. The setting screw was a lock/unlock to remove the rotor. 
    • I have that French tech sheet too, it is a little different than the English one (eg, it doesn't have the auto works diagram). BTW, it looks like you are looking up the case number in the 1979 ABC supplement. The 1974 ABC catalog does have the 3093 case. As you determined it takes the 1222-5 crystal.  When I serviced my President 'A' (which also takes that crystal), I was able to fit a 29.8 crystal from my DPA crystal assortment. Those are, in my opinion, a great deal. The assortment comes with 10 sizes each from 27.8mm to 32.4mm in 0.2 increments. I pretty much use them for any non-armored crystal that takes a high dome crystal. I think they no longer make them but Cousins has still has some in stock but when I bought them they were around $40 for the set and now they are around $100. Still, at 40 cents a crystal it's still a good deal. For the large driving wheel, I remember I once assembled the keyless/motion works first and when I placed the large driving wheel it was interfering with the setting wheel on the dial side as the teeth were not fully meshing and it wouldn't fully seat. If that isn't the issue I got nothing and am looking forward to see how you solve it 🙂
×
×
  • Create New...